Which pulse sequence parameter combination would yield the image with the most blurring?

Blur is driven mainly by voxel size and how k-space data is acquired during the echo train. A thicker slice increases partial-volume effects along the through-plane, so edges become less sharp. In fast spin-echo sequences, a longer echo train length means more echoes sample while signal decays, which blurs high-frequency details and smooths edges. The combination with a 5 mm slice thickness and a very long echo train length (16) emphasizes these effects, especially since the in-plane resolution isn’t extremely high. The other options improve sharpness by reducing voxel size (thinner slices or higher matrix), or by shortening the echo train (less T2-related blurring), so they produce crisper images.

Blur is driven mainly by voxel size and how k-space data is acquired during the echo train. A thicker slice increases partial-volume effects along the through-plane, so edges become less sharp. In fast spin-echo sequences, a longer echo train length means more echoes sample while signal decays, which blurs high-frequency details and smooths edges.

The combination with a 5 mm slice thickness and a very long echo train length (16) emphasizes these effects, especially since the in-plane resolution isn’t extremely high. The other options improve sharpness by reducing voxel size (thinner slices or higher matrix), or by shortening the echo train (less T2-related blurring), so they produce crisper images.